Regarding the values you found, I can't recall why I wrote this script. It was likely when we converted the PID controller to run on actuators instead of joints. In the case of the PSM, there is no coupling matrix in the script so what you get is the max actuator torques, i.e. torques at the sterile adapter. These are not the same as the joint maximum torques. Each instrument has its own coupling matrix (see files in share/tool/*.json). The cap on the maximum effort sent to motors is performed at the lowest level, i.e. in sawRobotIO1394. See https://github.com/jhu-saw/sawRobotIO1394/blob/d5e652d624f515ccfe534d3aa4a0141a5c6c40eb/components/code/mtsRobot1394.cpp#L1577 The max values are pulled from the sawRobotIO XML configurations files and are provided in amperes. These were provided by Intuitive in the dVRK User Guide. Regarding a sustained use at high torques, I don't really know if that could hurt either the instrument or the arm itself. Are you considering fast changes or something like holding an object for a long period? |
Beta Was this translation helpful? Give feedback.
-
Hi all,
I am wondering what the maximum allowed torque is for the last four actuators of the PSM (Classic da Vinci). I noticed this computation in compute-ftmax.py. This final result (0.3436) is in Newton meter, is that correct? Is this the actual (safe) amount of torque that the 4 actuators of the PSM can reach in practice? Could the robot be damaged if the actuators are kept working at close to this value for a long time (e.g., 1 hour)?